a) Cabinet body is made up of 1.2mm cold-rolled steel with anti-bacteria powder coating. The

structural strength and stability are enhanced.

b) Work area is made up of 304 stainless steel, which provides corrosion resistance and attractive

appearance.

c) Base stand is made up of metal with anti-bacteria powder coating.

d) Control panel adopts soft touch buttons, which provides attractive appearance and easy operation..

2.3 Instructions for Operation

2.3.1 Normal Operation Notice

a) Making sure input voltage is correct and stable. The rated load of main power socket should be

higher than cabinet consumption. Plug must be well grounded. If socket can not plug in, please

install a grounding type power socket by a electrician.

b) Moving principles of different samples inside cabinet: When two or more samples need to be

moved in, be sure that low-polluting samples should be moved first then high-polluting samples.

Movement of items should also follow the principles of moving slowly and steadily.

c) The weight of items placed in the cabinet should not be more than 23Kg/25

×

25cm2

d) Avoiding vibration: avoid using vibration equipment (eg centrifuges, vortex oscillator, etc.) inside

the cabinet. The contamination might drop from the HEPA filter and affect the samples and the

operator

e) NO FLAME: An open flame would create turbulence which disrupts the pattern of HEPA-filtered

air supplied to the work surface. If sterilization is required during the experiment, infrared sterilizer

is highly recommended.

f) HEPA filter life: With the usage time increasing, dust and bacteria accumulate inside HEPA filter.

Filter Resistance is getting bigger, when it reaches the maximum point, the speed requirements can’t

be met. Then need contact BIOBASE service department to get a new one. The used filter should be

processed as medical waste.

g) Air duct has been sealed strictly. Please do NOT remove or loose the screws of those parts. To

contact service personnel for special technical requirements

h) The maximum storage period is one year. A performance inspection should be done if the storage

period exceeds one year.

Serious declaration:

NOTE: BIOBASE WILL NOT BE RESPONSIBLE FOR ANY RISK OR DAMAGE

ARISING FAILURE DURING USAGE BY UNTRAINED OPERATOR AND

MISHANDLING THE LAF. THE OPERATOR OF THE LAMINAR FLOW CABINET

SHOULD BE TRAINED ENOUGH BEFORE USING THE LAF.

2.3.2 Operation Process

a) Connect the power

b) Press relevant button (please reference to 2.2.3 about button function and operation). Check if the

buttons in accordance to operation results. If the equipment is function normally, air velocity, UV

lamp and LED lamp according to technical parameters.

c) The cabinet should be sterilized by UV lamp for at least 30 minutes with the window fully closed

before any experiment.

Summary of Contents for BBS-DDC

Page 1: ...Vertical Laminar Flow Cabinet BBS DDC BBS SDC BBS DSC BBS SSC User Manual BIOBASE GROUP Version 2022 01 ...

Page 2: ...structions and Warranty carefully before operating this unit to assure proper operation After reading these documents be sure to store them securely together with the Warranty within reach for future reference Warning Before operating the unit be sure to read carefully and fully understand important warnings in the operating instructions ...

Page 3: ...rotected Object 14 2 1 4Technical Parameters 14 1 Vibration Amplitude 16 2 Illumination 16 3 Electrical Characteristic 16 2 2 Product Structure 17 2 2 1 Structural Composition of BBS DDC BBS SDC Picture 8 17 2 2 2 Structural composition of BBS DSC BBS SSC Picture 9 18 2 2 3 Structure Introduction 18 2 3 Instructions for Operation 20 2 3 1 Normal Operation Notice 20 2 3 2 Operation Process 20 2 4 D...

Page 4: ...acement parts list For parts purchase 22 2 6 Wiring Diagram Picture 12 24 3 Trouble Shooting and Solution 25 3 1 Common faults and solution 25 3 2 Replacement of Fuse 26 3 3 Replacement of LED Lamp 27 3 4 Replacement of UV Lamp 27 3 5 Label Description 28 4 Warranty 28 ...

Page 5: ... packing box is in good condition If the packing box is damaged please take photos 1 1 Unpacking Choose proper tools and unpacking method as shown in the below picture For wooden box 1 Method 1 Necessary tools for unpacking Electric drill with hexagon dead M8 Picture 1 ...

Page 6: ...nstrates quick unpacking procedures Picture 3 Remove the screws as shown in the below diagram then remove the wooden sheet to right and left of the wooden box Picture 3 For carton box Using scissor to cut the packing tape take off the package cover then move up the box body ...

Page 7: ...se protector 10A Packing bag behind the cabinet 1 pc UV Lamp T6 18W Paper package behind the cabinet 1 pc User Manual Paper package behind the cabinet 1set Inspection report Paper package behind the cabinet 1 pc Certificate of quality Paper package behind the cabinet 1 pc Warranty card Paper package behind the cabinet 1 pc Hexagon socket head cap screws M10 55 Paper package behind the cabinet 3set...

Page 8: ...ge behind the cabinet 1 pc Hexagon socket head cap screws M10 55 Paper package behind the cabinet 3 sets flat washer 10 spring washer 10 Paper package behind the cabinet 3 sets Hexagon wrench Paper package behind the cabinet 1 set BBS DSC Packing list Items Position Quantity Main body Inside the packing box 1 unit Base stand Paper package behind the cabinet 1 pc Power cord Packing bag behind the c...

Page 9: ...ead cap screws M10 55 Paper package behind the cabinet 5 sets flat washer 10 spring washer 10 Paper package behind the cabinet 5 sets Hexagon wrench Paper package behind the cabinet 1 set 1 3 Installation Conditions and Operating Environment Laminar flow cabinet should be placed in a position where the airflow can be protected Laminar flow cabinet should not be installed opposite to door or window...

Page 10: ...ysical damage or scratch the surface of main body during transit Make sure there is no damage scratch deformation or foreign bodies c Carefully check the accessories and material according to packing list in the manual d Before breaking the package move the entire equipment to the place where it is going to be installed NOTE Direct installation is completed for base stand part at the time of trans...

Page 11: ...both side of base stand Foot height can be adjusted Clockwise rotation of feet when feet height is less than casters the cabinet can be or base stand moved anticlockwise rotation of feet when feet height is greater than casters then play a fixed role in preventing the cabinet or base stand from moving NOTE If there are any questions please contact the Engineer for process debugging f Connect base ...

Page 12: ...11 a BBS DDC BBS SDC Hexagon socket head screws M10 55 Flat washers 10 Washers 10 Positioning bolt for installing cabinet ...

Page 13: ...the installation hole using M10 55 Inner six angle cylinder head bolt flat washers 10 and spring washers 10 bottom up through the base stand and side plate and fastened firmly g After the above steps slightly moving the cabinet to the proper position remove the power line power line inspection is intact Positioning bolt for installing cabinet Hexagon socket head screws M10 55 Flat washers 10 Washe...

Page 14: ...n check the follows items with power on Checking Items Normal working status Fan Runs normally LED Lamp Lamp lights up after pressing button UV Lamp Lamp lights up after pressing button Screen Buttons All buttons work effectively NOTE If there is any problem please contact the engineer for process debugging Debugging methods is in the after sales service manual ...

Page 15: ... pressure in order to prevent the infiltration of contaminated room air This model is vertical laminar flow cabinet 2 1 2 Operating Principle Air flow Pattern Picture 7 Picture 7 Air flow pattern 2 1 3 Protected Object Sample protection only 2 1 4Technical Parameters Model Parameters BBS DDC BBS SDC External Size W D H 1040 615 1770 mm 1440 615 1770 mm Working Zone Size W D H 940 540 545 mm 1340 5...

Page 16: ...t 2 Gross Weight 150kg 174kg Package Size W D H 1190 890 1340mm 1590 890 1330mm Model Parameters BBS DSC BBS SSC External Size W D H 1040 660 1770 mm 1440 660 1770 mm Working Zone Size W D H 940 560 545 mm 1340 560 545 mm Work Surface Height 750mm Display LED Display Airflow Velocity Average of 0 3 0 5m s Material Main Body Cold rolled steel with anti bacteria powder c oating Work Table 304 stainl...

Page 17: ...g 174kg Package Size W D H 1190 890 1340mm 1590 890 1340mm NOTE the company reserves all the rights to changes in product design if there are any design change we will not inform in advance 1 Vibration Amplitude The net vibration amplitude between 10Hz and 10KHz is less than 5μm rms 2 Illumination Average illumination is not less than 350 lx 3 Electrical Characteristic The voltage increases to 139...

Page 18: ...ture 2 2 1 Structural Composition of BBS DDC BBS SDC Picture 8 1 Control Panel 2 Power Socket 3 UV Lamp 4 LED Lamp 5 Front Window 6 Handle 7 Work Table 8 Base Stand 9 Leveling Feet 10 Caster Picture 8 4 1 7 9 10 8 6 5 3 2 ...

Page 19: ...filtration is the core system of laminar flow cabinet It consists of blower and filter The main function of air filtration system is to transfer filtered air to work area ensure the flow velocity and keep cleanness of work area 3 UV lamp The UV lamp located at the top of the work zone effectively disinfects the entire work zone by emitting a wavelength of 253 7nm 4 LED lamp Laminar flow cabinet is...

Page 20: ...er during working Air velocity control During working button to increase air velocity button to decrease air velocity 6 Fuse protector The cabinet is equipped with main power fuse For double pane window it is located beside the power cord s outlet For single glass they are both located at the back of the power cord s outlet Fuse label is corresponding to the relevant specifications Please refer to...

Page 21: ...e NO FLAME An open flame would create turbulence which disrupts the pattern of HEPA filtered air supplied to the work surface If sterilization is required during the experiment infrared sterilizer is highly recommended f HEPA filter life With the usage time increasing dust and bacteria accumulate inside HEPA filter Filter Resistance is getting bigger when it reaches the maximum point the speed req...

Page 22: ...which has been soaked with clean hot or warm water At the end wipe the entire surface with a dry cotton cloth or towel rapidly For the contaminated or dirty work surface and sump use medical alcohol or other disinfectant to wipe NOTE Disinfectants used for wiping should not damage the 304 stainless steel 2 4 2 Clean the external surface and front window Use a piece of soft cotton cloth or towel wi...

Page 23: ... or other corrosive gases Storage period shall not exceed one year Laminar flow cabinet stored for more than one year needs to be unpacked and checked before selling and using Only the tested and qualified Laminar Flow Cabinet could be sold 2 5 Replacement parts list For parts purchase BBS DDC replacement parts list Number Name Specification BE 01 Fuse 10A BE 02 Lamp holder T8 LG13 01A BE 03 UV La...

Page 24: ...stant force spring 405S nylon sleeve nylon sleeve baffle BBS DSC replacement parts list Number Name Specification BF 01 Fuse 10A BF 02 Lamp holder T8 LG13 01A BF 03 UV Lamp T6 18W BF 04 LED T5 Supporting Frame T5 12W BF 05 UV lamp ballast 1 TL8 18W BF 06 High efficient filter 900 450 50mm BF 07 Fan FS133Q BF 08 Control panel Laminar flow cabinet control panel Ten light pillar BF 09 Glass 960 580 5...

Page 25: ...ing 405L nylon sleeve nylon sleeve baffle BK 11 Constant force spring Constant force spring 405S nylon sleeve nylon sleeve baffle 2 6 Wiring Diagram Picture 12 1 2 3 4 A B C D 4 3 2 1 D C B A fan FUSE1 10A L N mcu brown Φ1 brown Φ1 blue Φ1 0 b l u e Φ 1 transformer l i g h t u v L fan y e l l o w Φ 1 b l u e Φ 1 blue Φ1 12V 0 3A UV LAMP U V B A L L A S T BLACK Φ1 WHITE Φ0 5 BLACK Φ0 5 WHITE Φ0 5 B...

Page 26: ...eck the blower and LED is turn off or not Lamp holder Connect the tube and lamp holder tightly Circuit Check the circuit UV lamp Replace the UV lamp Ballast Replace the ballast Control panel Replace the control panel Button fail to work Control panel Make sure the power is well connected and the fuse is in good condition Check if the button is broken Make sure the connecting wire is well connected...

Page 27: ...g and repair of this equipment only could be undertaken by trained and recognized technicians 4 Please contact Biobase or our agent technical department or agent to order required component or part The model number and the serial number of purchased cabinet need to be indicated 3 2 Replacement of Fuse For single side operation laminar flow cabinet power input socket is located at the back for doub...

Page 28: ...re 13 Picture 14 3 4 Replacement of UV Lamp UV lamp should be replaced regularly according to the frequency of use when using UV lamps reach to the time of 600 hours we recommend to replace the lamp UV intensity test card can be used to confirm if need to change UV lamp Please turn off the power Screw the lamp to take down from holder then replace with a new one Please reference to picture 15 LED ...

Page 29: ...d LED lamp fuse 4 1 2 Biobase would not be liable for any repair of damage caused by improper operation 4 1 3 If the warranty has been expired Biobase would still responsible for repair with relative charges 4 1 4 Life time of laminar flow cabinet is 8 years from production date on the label 4 1 5 Biobase would provide equipment drawings and necessary technical data for maintenance companies or pe...

Page 30: ...inan City Shandong Province China Tel 86 531 81219803 01 Fax 86 531 81219804 Inquiry export biobase com Complaints customer_support biobase cc After sales service service_sd biobase cc service_ivd biobase cc Web www biobase cc www meihuatrade com www biobase com ...

Reviews: