29
Note (1)
Above 1000V, a repetitive beep indicates that the voltage being
measured is greater than the safety voltage for which the device is guaranteed.
4.2.2
AC voltage measurement
Measurement
range
0.15V to
99.99V
100.0V to
999.9V
1000V RMS
1400V peak (1)
Specified
measurement
range (2)
0 to 100% of the measurement range
Uncertainties
from 0.15V to 9.99V
± (1% R + 10 pt)
from 10.00V to
99.99V
± (1% R +3 pt)
± (1% R +3 pt)
Resolution
0.01 V
0.1 V
1 V
Input impedance
10 M
Ω
Note (1) Above 1000V, a repetitive beep indicates that the voltage being measured
is greater than the safety voltage for which the device is guaranteed.
Bandwidth in AC = 3 kHz
Note (2) Any value between zero and the min. threshold of the measurement range
(0.15V) is forced to "----" on the display
Specific characteristics in MAX/MIN mode
(from 10Hz to 1kHz in AC, from
0.30V) :
•
Uncertainties: add 1% L to the values of the table above.
•
Capture time of the extrema: approximately 100ms.
4.2.3
AC current measurement
Measurement
range (2)
0.25 A to
99.99 A
100.0 A to
999.9 A
1000 A (1)
Specified
measurement
range
0 to 100% of the measurement range
Uncertainties
± (1% R + 10 pt)
± (1% R +3 pt)
Resolution
0.01 A
0.1 A
1 A
Summary of Contents for F401
Page 1: ...F401 Clamp multimeter GB User s manual...