57
- Above 1000V (DC or RMS), a repetitive beep indicates that the voltage
being measured is greater than the safety voltage for which the device is
guaranteed.
-
Bandwidth in AC = 3 kHz
Note (2)
- Any value between zero and the min. threshold of the measurement
range (0.15V) is forced to "----" on the display.
- Specific characteristics in MAX/MIN mode
(from 10 Hz to 1 kHz in AC and
AC+DC, and since 0,30V):
•
Uncertainties: add 1% L to the values in the tables above.
•
Capture time of the extrema: approximately 100ms.
- Specific characteristics in PEAK mode
(from 10 Hz to 1 kHz in AC and
AC+DC):
•
Uncertainties: add 1.5% L to the values in the tables above.
•
PEAK capture time: 1ms min. to 1.5ms max.
4.2.4
DC current measurement
Measurement range
0.00A to
99.99A
100.0A to
999.9A
1000A to 1500A
(1)
Specified
measurement range
0 to 100% of the measurement range
Uncertainties (2)
(zero corrected)
± (1% R+10 pt) ± (1% R +3 pt)
± (1,5% R +3 pt)
Resolution
0.01A
0.1A
1A
Note (1) -
The display indicates “+OL" above 1500A .
Note (2) - The residual current at zero depends on the remanence; It can be
corrected by the “DC zero” function of the HOLD key.
4.2.5
AC current measurement
Measurement
range (2)
0.25 A to
99.99 A
100.0 A to
999.9 A
1000 A (1)
Specified
measurement
range
0 to 100% of the measurement range
Uncertainties
± (1% R + 10 pt)
± (1% R +3 pt)
Resolution
0.01A
0.1A
1A