58
Note (1) - The display indicates "OL" above 1500A (in PEAK mode). The "-
"
and "+" signs are not managed.
-
Bandwidth in AC = 2 kHz
Note (2) - Any value between zero and the min. threshold of the measurement
range (0.25A) is forced to “----“ on the display.
4.2.6
AC+DC intensity measurement
Measurement
range (2)
0.25A to
99.99A
100.0A to
999.9A
AC: 1000A
DC or PEAK: 1000A to
1500A (1)
Specified
measurement
range
0 to 100% of the measurement range
Uncertainties
(2)
(zero corrected)
± (1% R+10 pt) ± (1% R +3pt) ± (1,5% R +3pt)
Resolution
0.01A
0.1A
1A
Note (1) - The display indicates "+OL" above 1500A (in PEAK mode). The "-" and
"+" signs are not managed.
-
Bandwidth in AC = 2 kHz
Note (2) - In AC, any value between zero and the min. threshold of the
measurement range (0.25A) is forced to "----" on the display.
- Specific characteristics in MAX/MIN mode
(from 10 Hz to 1 kHz in AC and
AC+DC, and since 0,30A):
•
Uncertainties: add 1% R to the values in the tables above.
•
Capture time of the extrema: approximately 100ms.
- Specific characteristics in PEAK mode
(from 10 Hz to 1 kHz in AC and
AC+DC):
•
Uncertainties: add 1.5% L to the values in the tables above.
•
PEAK capture time: 1ms min. to 1.5ms max.
4.2.7
True-Inrush measurement
Measurement range
10 A to 1000 A AC
10 A to 1500 A DC
Specified measurement range
0 to 100% of the measurement range
Uncertainties
± (5% R + 5 pt)
Resolution
1 A
Summary of Contents for F407