64
4.2.18 Measurement of reactive AC+DC power
Measurement
range (2) (4)
5 var to
9999 var
10,00kvar
to
99,99kvar
100,0kvar
to
999,9kvar
1000kvar to
1500kvar
(1)
Specified
measurement
range
1 to 100% of the
measurement range
0 to 100% of the measurement range
Uncertainties
(3) (8)
until 1000A
± (2% R +10 pt)
from 1000A to 1500A
± (2.5% R +10 pt)
until 1000A
± (2% R +3 pt)
from 1000A to 1500A
± (2.5% R +3 pt)
Resolution
1 var
10 var
100 var
1 kvar
Note (1) - Display of O.L above 1500 kvar in single-phase (1000 V x 1500 A).
- Bandwidth in AC in voltage = 3 kHz, in current = 2 kHz
Notes (2), (3) , (4), 5, 6 and (8) of the previous § apply.
- Specific characteristics in MAX/MIN mode in power
(from 10Hz to 1kHz in AC
and AC+DC):
•
Uncertainties: add 1 %R to the values in the tables above.
•
Capture time: approximately 100ms
4.2.19 Calculation of the power factor (PF)
Measurement range (1)
0.00 to +1.00
Specified measurement range
0 to 50% of the
measurement range
50 to 100% of the
measurement range
Uncertainties (7)
± (3% R +3 pt)
± (2% R +3 pt)
Resolution
0.01
Note (1) - If one of the terms in the calculation of the power factor is displayed as
"OL", or forced to zero, the display of the power factor is an indeterminate
value "----".
Note (7) of the previous § apply.
Remark : The PF is always positive
- Specific characteristics in MAX/MIN mode
(from 10Hz to 1kHz):
•
Uncertainties: add 1 %R to the values in the tables above.
•
Capture time: approximately 100ms.