63
4.2.16 Measurement of apparent AC+DC power
Measurement
range(2) (4)
5 VA to
9999 VA
10,00 kVA
to
99,99 kVA
100,0 kVA
to
999,9 kVA
1000 kVA to
1500 kVA (1)
Specified
measurement
range
1 à 100% of the
measurement range
0 to 100% of the measurement range
Uncertainties
(3)
until 1 000 A
± (2% R +10 pt)
from 1000 A to 1500 A
± (2,5% R +10 pt)
until 1 000 A
± (2% R +3 pt)
from 1000 A to 1500 A
± (2,5% R +3 pt)
Resolution
1 VA
10 VA
100 VA
1 000 VA
Note (1) - Display of O.L above 1500 kVA in single-phase (1000 V x 1500 A).
- Bandwidth in AC in voltage = 3 kHz, in current = 2 kHz
Notes (2), (3) and (4) of the previous § apply.
4.2.17 Measurement of reactive AC power
Measurement
range (2) (4)
5 var to
9999 var
10,00 kvar
to
99,99 kvar
100,0 kvar to
999,9 kvar
1000 kvar (1)
Specified
measurement
range
1 to 100% of the
measurement range
0 to 100% of the measurement range
Uncertainties
(3) (8)
± (2% R +10 pt)
± (2% R +3 pt)
Resolution
1 var
10 var
100 var
1 kvar
Note (1) - Bandwidth in AC in voltage = 3 kHz, in current = 2 kHz
Notes (2), (3) and (4) of the previous § apply.
Note 5 - In single-phase, the sign of the reactive power is determined by the
phase lead or lag between the U and I signs, while in balanced three-
phase, it is determined by the calculation on the samples.
Note 6 - Signs of reactive powers according to the four-quadrant rule (§4.2.12):
Quadrant 1 : Reactive power Q sign +
Quadrant 2 : Reactive power Q sign +
Quadrant 3 : Reactive power Q sign -
Quadrant 4 : Reactive powerQ
sign -
Note (8) - Measurement stabilization ~8 sec